Sterilization and Disinfection

Home Forums MISCELLANEOUS POSTS MISCELLANEOUS POSTS Sterilization and Disinfection Sterilization and Disinfection

#14603
drmithiladrmithila
Offline
Registered On: 14/05/2011
Topics: 242
Replies: 579
Has thanked: 0 times
Been thanked: 0 times

STERILIZATION OR DISINFECTION OF INSTRUMENTS Indications for Sterilization or Disinfection of Dental Instruments

As with other medical and surgical instruments, dental instruments are classified into three categories — critical, semicritical, or noncritical — depending on their risk of transmitting infection and the need to sterilize them between uses (9,37-40). Each dental practice should classify all instruments as follows:

Critical. Surgical and other instruments used to penetrate soft tissue or bone are classified as critical and should be sterilized after each use. These devices include forceps, scalpels, bone chisels, scalers, and burs.

Semicritical. Instruments such as mirrors and amalgam condensers that do not penetrate soft tissues or bone but contact oral tissues are classified as semicritical. These devices should be sterilized after each use. If, however, sterilization is not feasible because the instrument will be damaged by heat, the instrument should receive, at a minimum, high-level disinfection.

Noncritical. Instruments or medical devices such as external components of x-ray heads that come into contact only with intact skin are classified as noncritical. Because these noncritical surfaces have a relatively low risk of transmitting infection, they may be reprocessed between patients with intermediate-level or low-level disinfection (see Cleaning and Disinfection of Dental Unit and Environmental Surfaces) or detergent and water washing, depending on the nature of the surface and the degree and nature of the contamination (9,38). Methods of Sterilization or Disinfection of Dental Instruments

Before sterilization or high-level disinfection, instruments should be cleaned thoroughly to remove debris. Persons involved in cleaning and reprocessing instruments should wear heavy-duty (reusable utility) gloves to lessen the risk of hand injuries. Placing instruments into a container of water or disinfectant/detergent as soon as possible after use will prevent drying of patient material and make cleaning easier and more efficient. Cleaning may be accomplished by thorough scrubbing with soap and water or a detergent solution, or with a mechanical device (e.g., an ultrasonic cleaner). The use of covered ultrasonic cleaners, when possible, is recommended to increase efficiency of cleaning and to reduce handling of sharp instruments.

All critical and semicritical dental instruments that are heat stable should be sterilized routinely between uses by steam under pressure (autoclaving), dry heat, or chemical vapor, following the instructions of the manufacturers of the instruments and the sterilizers. Critical and semicritical instruments that will not be used immediately should be packaged before sterilization.

Proper functioning of sterilization cycles should be verified by the periodic use (at least weekly) of biologic indicators (i.e., spore tests) (3,9). Heat-sensitive chemical indicators (e.g., those that change color after exposure to heat) alone do not ensure adequacy of a sterilization cycle but may be used on the outside of each pack to identify packs that have been processed through the heating cycle. A simple and inexpensive method to confirm heat penetration to all instruments during each cycle is the use of a chemical indicator inside and in the center of either a load of unwrapped instruments or in each multiple instrument pack (41); this procedure is recommended for use in all dental practices. Instructions provided by the manufacturers of medical/dental instruments and sterilization devices should be followed closely.

In all dental and other health-care settings, indications for the use of liquid chemical germicides to sterilize instruments (i.e., “cold sterilization”) are limited. For heat-sensitive instruments, this procedure may require up to 10 hours of exposure to a liquid chemical agent registered with the U.S. Environmental Protection Agency (EPA) as a “sterilant/disinfectant.” This sterilization process should be followed by aseptic rinsing with sterile water, drying, and, if the instrument is not used immediately, placement in a sterile container.

EPA-registered “sterilant/disinfectant” chemicals are used to attain high-level disinfection of heat-sensitive semicritical medical and dental instruments. The product manufacturers’ directions regarding appropriate concentration and exposure time should be followed closely. The EPA classification of the liquid chemical agent (i.e., “sterilant/disinfectant”) will be shown on the chemical label. Liquid chemical agents that are less potent than the “sterilant/disinfectant” category are not appropriate for reprocessing critical or semicritical dental instruments.

CLEANING AND DISINFECTION OF DENTAL UNIT AND ENVIRONMENTAL SURFACES

After treatment of each patient and at the completion of daily work activities, countertops and dental unit surfaces that may have become contaminated with patient material should be cleaned with disposable toweling, using an appropriate cleaning agent and water as necessary. Surfaces then should be disinfected with a suitable chemical germicide.

A chemical germicide registered with the EPA as a “hospital disinfectant” and labeled for “tuberculocidal” (i.e., mycobactericidal) activity is recommended for disinfecting surfaces that have been soiled with patient material. These intermediate-level disinfectants include phenolics, iodophors, and chlorine-containing compounds. Because mycobacteria are among the most resistant groups of microorganisms, germicides effective against mycobacteria should be effective against many other bacterial and viral pathogens (9,38-40,42). A fresh solution of sodium hypochlorite (household bleach) prepared daily is an inexpensive and effective intermediate-level germicide. Concentrations ranging from 500 to 800 ppm of chlorine (a 1:100 dilution of bleach and tap water or 1/4 cup of bleach to 1 gallon of water) are effective on environmental surfaces that have been cleaned of visible contamination. Caution should be exercised, since chlorine solutions are corrosive to metals, especially aluminum.

Low-level disinfectants — EPA-registered “hospital disinfectants” that are not labeled for “tuberculocidal” activity (e.g., quaternary ammonium compounds) — are appropriate for general housekeeping purposes such as cleaning floors, walls, and other housekeeping surfaces. Intermediate- and low-level disinfectants are not recommended for reprocessing critical or semicritical dental instruments.

DISINFECTION AND THE DENTAL LABORATORY

Laboratory materials and other items that have been used in the mouth (e.g., impressions, bite registrations, fixed and removable prostheses, orthodontic appliances) should be cleaned and disinfected before being manipulated in the laboratory, whether an on-site or remote location (43). These items also should be cleaned and disinfected after being manipulated in the dental laboratory and before placement in the patient’s mouth (2). Because of the increasing variety of dental materials used intraorally, DHCWs are advised to consult with manufacturers regarding the stability of specific materials relative to disinfection procedures. A chemical germicide having at least an intermediate level of activity (i.e., “tuberculocidal hospital disinfectant”) is appropriate for such disinfection. Communication between dental office and dental laboratory personnel regarding the handling and decontamination of supplies and materials is important.