USE AND CARE OF HANDPIECES, ANTIRETRACTION VALVES, AND OTHER INTRAORAL DENTAL DEVICES ATTACHED TO AIR AND WATER LINES OF DENTAL UNITS

Routine between-patient use of a heating process capable of sterilization (i.e., steam under pressure {autoclaving}, dry heat, or heat/chemical vapor) is recommended for all high-speed dental handpieces, low-speed handpiece components used intraorally, and reusable prophylaxis angles. Manufacturers’ instructions for cleaning, lubrication, and sterilization procedures should be followed closely to ensure both the effectiveness of the sterilization process and the longevity of these instruments. According to manufacturers, virtually all high-speed and low-speed handpieces in production today are heat tolerant, and most heat-sensitive models manufactured earlier can be retrofitted with heat-stable components.

Internal surfaces of high-speed handpieces, low-speed handpiece components, and prophylaxis angles may become contaminated with patient material during use. This retained patient material then may be expelled intraorally during subsequent uses (44-46). Restricted physical access — particularly to internal surfaces of these instruments — limits cleaning and disinfection or sterilization with liquid chemical germicides. Surface disinfection by wiping or soaking in liquid chemical germicides is not an acceptable method for reprocessing high-speed handpieces, low-speed handpiece components used intraorally, or reusable prophylaxis angles.

Because retraction valves in dental unit water lines may cause aspiration of patient material back into the handpiece and water lines, antiretraction valves (one-way flow check valves) should be installed to prevent fluid aspiration and to reduce the risk of transfer of potentially infective material (47). Routine maintenance of antiretraction valves is necessary to ensure effectiveness; the dental unit manufacturer should be consulted to establish an appropriate maintenance routine.

High-speed handpieces should be run to discharge water and air for a minimum of 20-30 seconds after use on each patient. This procedure is intended to aid in physically flushing out patient material that may have entered the turbine and air or water lines (46). Use of an enclosed container or high-velocity evacuation should be considered to minimize the spread of spray, spatter, and aerosols generated during discharge procedures. Additionally, there is evidence that overnight or weekend microbial accumulation in water lines can be reduced substantially by removing the handpiece and allowing water lines to run and to discharge water for several minutes at the beginning of each clinic day (48). Sterile saline or sterile water should be used as a coolant/irrigator when surgical procedures involving the cutting of bone are performed.

Other reusable intraoral instruments attached to, but removable from, the dental unit air or water lines — such as ultrasonic scaler tips and component parts and air/water syringe tips — should be cleaned and sterilized after treatment of each patient in the same manner as handpieces, which was described previously. Manufacturers’ directions for reprocessing should be followed to ensure effectiveness of the process as well as longevity of the instruments.

Some dental instruments have components that are heat sensitive or are permanently attached to dental unit water lines. Some items may not enter the patient’s oral cavity, but are likely to become contaminated with oral fluids during treatment procedures, including, for example, handles or dental unit attachments of saliva ejectors, high-speed air evacuators, and air/water syringes. These components should be covered with impervious barriers that are changed after each use or, if the surface permits, carefully cleaned and then treated with a chemical germicide having at least an intermediate level of activity. As with high-speed dental handpieces, water lines to all instruments should be flushed thoroughly after the treatment of each patient; flushing at the beginning of each clinic day also is recommended.